Question:
Measuring time not valid showing in the audit trail and the device time is always the same.
Possible Root Cause:
The RMS-MLOG antenna creates short circuits when it touches the PCB. Due to this fact, a redesign of the antenna was carried out where the antenna shape was adapted.
The RMS-MLOG battery contacts have been known to oxidize. A cleaning of the contacts may be required in some cases.
Measuring time not valid showing in the audit trail and the device time is always the same.
Possible Root Cause:
- The battery is empty.
- The antenna is touching the PCB.
- The battery contacts are oxidized.
- Change the battery. Please check the guidelines: https://service.rotronic.com/manual/...-battery.html?.
- Bend the antenna: https://service.rotronic.com/manual/...lled-on-t.html.
- Clean the battery contacts (simply by removing and adding the battery again).
The RMS-MLOG antenna creates short circuits when it touches the PCB. Due to this fact, a redesign of the antenna was carried out where the antenna shape was adapted.
The RMS-MLOG battery contacts have been known to oxidize. A cleaning of the contacts may be required in some cases.