    Question:
    Measuring time not valid showing in the audit trail and the device time is always the same.

    Possible Root Cause:
    1. The battery is empty.
    2. The antenna is touching the PCB.
    3. The battery contacts are oxidized.
    Solution:
    1. Change the battery. Please check the guidelines: https://service.rotronic.com/manual/...-battery.html?.
    2. Bend the antenna: https://service.rotronic.com/manual/...lled-on-t.html.
    3. Clean the battery contacts (simply by removing and adding the battery again).
    Additional Information:
    The RMS-MLOG antenna creates short circuits when it touches the PCB. Due to this fact, a redesign of the antenna was carried out where the antenna shape was adapted.
    The RMS-MLOG battery contacts have been known to oxidize. A cleaning of the contacts may be required in some cases.