If you replace the battery within the HL-1D will readings stored in memory be maintained? Also, will the calibration data also remain intact or will the logger require calibration after a battery replacement?
Announcement
Collapse
No announcement yet.
HL-1D Battery Replacement
Collapse
X
-
Hi, the stored data will remain in the memory if the battery will taken out.
Calibration data are not stored generally. After calibration a calibration report can be created.
But adjustment data will be stored within the memory. So these data might not be lost when the battery is out of the device
-
Hi Michael,
for my understanding, a calibration is a measurement comparing the HL-1D to another reference. This kind of measurement are possible using HW4. But no calibration data will be stored within the HL-1D.
For an adjustment it is different. An adjustment of the HL-1D is a correction of the measurement of the HL-1D. With an adjustment, the acquired data (an offset of humidity) will be stored within the HL-1D.
If you take the battery out now, the adjustment data (acquired data/offset of humidity) keep within the device. But calibration data are never stored within the HL-1D independend if you take the battery out or not.
I hope I could explain my post a bit better.
Comment
-
I think when replacing the battery in the HL-1D data recorder, the data stored in the memory will usually be maintained. However, this still depends on the type of battery and how the computer is designed.