  • Validation Function: RMS-WEB-P V1.3.1 and HL-NT3

    I am using the MS-WEB-P V1.3.1 and HL-NT3 Datalogger with Ethernet Docking Station and I need the validation function to test my alarms and E-Mails.

    The first step within the validation is to send a simulator value. The step never works: why?
    It always says error...

  • #2
    James
    Humidity and Temperature of the HL-NT can be set as fixed value. But you must use the HW4 software to configure the HL-NT with probes connected.
    Disable the HL-NT at the RMS, open HW4, add the HL-NT to HW4 and configure the probe by setting the wished fix values.
    Turn off the HW4 and enable the HL-NT within the RMS. Than the measured values appear as "Simulator" - please see Manual of the RMS Converter ( link: https://www.rotronic.com/de-ch/produ...wnload?id=1358 ) page 98.

    Why? - The HL-NT is a Rotronic device of the first generation and not an IOT Device with RMS standards. Not every RMS functionality is supported by the HL-NT. Please see pages 94ff of the same manual.