Question:
Calibration of NTC's (T10-0002) at -80°C require readings lower than -80°C. The lookup table of the T10-0002 is blocked at -80°C.
Possible Root Cause:
The lookup table of the RMS-MLOG-T10 is stored upon the EEPROM of the MLOG. The user must ensure that the latest update of the EEPROM has been carried out.
10/12/2019: The actual MLOG memory is limited additional lookup tables cannot be implemented.
Calibration of NTC's (T10-0002) at -80°C require readings lower than -80°C. The lookup table of the T10-0002 is blocked at -80°C.
Possible Root Cause:
- The T10-0002 is limited to -80°C due to the lookup table.
- Please use the T10-0005
- Input the nominal value and b constant into the device properties via the RMS software. The nominal value and b constant data can be found within the datasheet. Please be aware that the temperature range is limited when using this solution.
The lookup table of the RMS-MLOG-T10 is stored upon the EEPROM of the MLOG. The user must ensure that the latest update of the EEPROM has been carried out.
10/12/2019: The actual MLOG memory is limited additional lookup tables cannot be implemented.