  • Case 171

    Question:
    Calibration of NTC's (T10-0002) at -80°C require readings lower than -80°C. The lookup table of the T10-0002 is blocked at -80°C.

    Possible Root Cause:
    1. The T10-0002 is limited to -80°C due to the lookup table.
    Solution:
    1. Please use the T10-0005
    2. Input the nominal value and b constant into the device properties via the RMS software. The nominal value and b constant data can be found within the datasheet. Please be aware that the temperature range is limited when using this solution.
    Additional Information:
    The lookup table of the RMS-MLOG-T10 is stored upon the EEPROM of the MLOG. The user must ensure that the latest update of the EEPROM has been carried out.
    10/12/2019: The actual MLOG memory is limited additional lookup tables cannot be implemented.