Question:
Probe serial number shows 0 in the RMS software instead of the real serial number
Possible Root Cause:
Probe serial number shows 0 in the RMS software instead of the real serial number
Possible Root Cause:
- The RMS-LOG is not online.
- The serial number has not yet been communicated to the database.
- The wireless settings of the logger and the gateway many not correspond.
- Make sure that the logger is online.
- Use the replace function to replace the logger with the same logger.
- Adapt the gateway wireless settings.