  • Case 200

    Question:
    RMS-MLOG-T10-868 showing N/A when the T10-0001 is attached. When using the T10-0003 the values are shown correctly.

    Possible Root Cause:
    1. The T10-0001 is designed for low temperature applications. The T10-0001 lookup table is only setup from -200 to -90°C. When adding the R25 and B constants, the user can also only see values between -200 and -95°C. No matter what option is selected, at 23°C, the user won't get any results.
    Solution:
    1. Carry out a measurement at -90°C to get readings.
    Link:
    https://service.rotronic.com/manual/...0-868_915.html

    Additional Information:
    The lookup table for the T10's is stored on the EEPROM of the RMS-MLOG-T10-868/915 and not on the firmware. In the RMS software, the RMS-MLOG-T10-868/915 will only show the lookup tables that are programmed on the EEPROM independant of the firmware installed on the device.
    Today it is not possible to update the EEPROM to add the various lookup tables without intervention from Rotronic AG.