  • Case 262

    Question:
    RMS-CCD shows wrong values if you read out with modbus commmands.

    Possible Root Cause:
    Wrong calculation in the modbus command.

    Solution:
    Use one of this calculation in the modbus commands to get the right results:
    0…2000ppm: 15
    0…5000ppm: 6
    0…10000ppm: 3