Question:
Two measuring points are not linked in the graphic even thought the measurement data is available.
Solution:
When a timeout occurs, a "--" value is written in the database so that the data gap can be recognised. The "--" will not be deleted, even when the measurement data is downloaded. As of such the measuring points are not linked within the chart.