Case 234

Thread #725 | 2020-03-19 10:59 UTC | James | replies: 0
Post #1 | 2020-03-19 10:59 UTC | James
Question:
Measuring time not valid showing in the audit trail and the device time is always the same.

Possible Root Cause:
  1. The battery is empty.
  2. The antenna is touching the PCB.
  3. The battery contacts are oxidized.
Solution:
  1. Change the battery. Please check the guidelines: https://service.rotronic.com/manual/...-battery.html?.
  2. Bend the antenna: https://service.rotronic.com/manual/...lled-on-t.html.
  3. Clean the battery contacts (simply by removing and adding the battery again).
Additional Information:
The RMS-MLOG antenna creates short circuits when it touches the PCB. Due to this fact, a redesign of the antenna was carried out where the antenna shape was adapted.
The RMS-MLOG battery contacts have been known to oxidize. A cleaning of the contacts may be required in some cases.