Question:
RMS-CCD shows wrong values if you read out with modbus commmands.

Possible Root Cause:
Wrong calculation in the modbus command.

Solution:
Use one of this calculation in the modbus commands to get the right results:
0…2000ppm: 15
0…5000ppm: 6
0…10000ppm: 3