Rotronic RMS Forum

Alarm on virtual button

Thread #643 | 2020-01-28 07:34 UTC | James | replies: 3
Post #1 | 2020-01-28 07:34 UTC | James
When adding an alarm on a virtual button, the alarm can only be set at 0.01 or 0.99. Should any other value be added, it will not be saved!
Post #2 | 2020-01-28 08:29 UTC | Clemens.Duzy
Hi James,
I tried it out too. If you set values for the upper limit of e. g. 0.8 and save it, than the RMS rounds the limit from 0.8 to 0.99. Same for the lower limit. E. G. a value of 0.2 is rounded to 0.01. The limits are rounded and saved but not principally lost.
Post #3 | 2020-01-28 08:50 UTC | James
Thanks Clemens, but if I try to update 0.99 to 0.4 for example, even though it is written in the audit trail, it won't be shifted to 0.01.
Post #4 | 2020-01-28 09:47 UTC | Clemens.Duzy
Hi James, that's true. So the information within the audit trail might be improved. I will take it into the issue list.