Rotronic RMS Forum

Measure Solar Radiation in RMS

Article #989 | 2020-11-04 13:46 UTC | Clemens.Duzy
Hi RMS Community,

at the Rotronic location in Bassersdorf we have an experimental wether station that we implemented into the RMS.
Today I would like to show you how to integrate a very simple solar radiation measurement.
It could be used to detect any solar influence for climate measurements as temperature, humidity or for electrical measurements as power or efficiency of a photovoltaic system.
A very accurate but also expensive sensor would be a pyranometer.

An economic alternativ is a Silicon Solar Cell.


Measurement principle
A solar cell supplies a current that depends on the solar radiation lineary.
To measure the current, a ohmic load is to be connected to the solar cell. Now the voltage over the load is to be measured as a rate of the solar radiation.

Voltage-Current-Diagram of a silicon solar cell

solar_cell_diagram.png

Setup
The setup is pretty simple, an ohmic resistor and a MiniLogger with a standard ADC interface of 0-10V (RMS-MADC-868-V) is to be used.

Circuit diagram of the setup

circuit.png



How to rate the ohmic load?
Looking into the data sheet of the solar cell there are two electric variables:
  • Rated voltage = 6.0V
  • Rated current = 0.12A
These represents the voltage and current at the most power point MPP at 1000W/m2 solar radiation.

The next diagram shows, how to interpret


Rating_of_a_resistor_load.png


The resistance of the load at the maximum power point MPP is 6.0V / 0.12V = 50Ohm.
The next standadized resistance would be 56Ohm.
The voltage over the load is
  • app. 5.6V at 1000W/m2 solar radiation
  • app. 2.8V at 500W/m2 solar radiation and so on.
The resisitor should support also the power of 0.72W on the MPP. So using a resistor of >1W would be sufficient.


Experimental setup

setup.png


Input scale of the MiniLogger
Please click on Tools -> Setup -> Devices -> Choose the MiniLogger -> Click on Show Properties:
  • The Measurement range low should be set to 0 (0W solar radiation at night).
  • The Measurement range high should be set to the radiation power expected at a signal of 10V. As the setup never supplies a voltage much higher than 5.6V, the theoretical max. scaling might be caluclated: (1000W/5.6V)*10V=1786W.
Click on Apply and you measure solar radiation in W/m2.


How accurate is this measurement?
There are some influences on the measurement as:
  • Temperature -> small influence on the voltage
  • Variation of Air Mass (AM) -> a rate represents the earth athmosphere homoginious dlyistributed (for solar cells only in laboratory reproducable)
  • Frequency sprectrum of the sun light -> Silicon detects only a part of the sun light spectrum. CO2 or H2O in the athmosphere e. g. can influence solar radation landing on the earth surface at higher frequencies. These influences cannot be measured by silicon.
  • Voltage over the load -> The load might be chosen higher than 50Ohm in the example above to detect the linear light dependency of the current. If the load or the current differes from datasheet for several further reason, than you have further influences on the measurement.
  • Dust and polution on the solar cell surface over the time.
A really easy way to adjust this setup is to get a really sunny day in April or May (nothern hemishpere). The maximum solar radiation is app. 1000 to perhaps. 1050W/m2. Scale your input this max. measurement and you got it. Than your measurement is a good indicator for the sun power in your application.


Diagram of the solar radiation over 12 hours

Diagram_solar_radiation.png

No matching content in this article.